You got me wondering
@damon about those Klein meters. I have a MM420 that I carry in the truck. This carries the same specifications for accuracy and resolution as yours. Adds a few more features but I suspect is going to be similar on the core.
Compared it against a Fluke 87 Series I from about 1994. It has almost identical specs to the Klein of 2023 and has over the years been NIST-traceable calibrated, although it is not currently since I had to break the seal for a battery.
This is just a 12V pack (measures 12.4V) feeding a 680Ω resistor, so should be 18.24mA on paper. Both agree very well within all the tolerances and limits of this junk box stuff. I also flipped polarity on them which did not confuse the Fluke (which I knew didn't care) nor the Klein. So that speculation on my part is completely bogus.
I see no reason to suspect your DMM isn't accurate, mine at least seems like a completely fine meter.
Using 10A ranges.
Using mA ranges.
And it later occurred to me that "What if the current
is actually 250mA, what might the meter show on the 200mA range?" It should recognize that and as long as you don't blow the fuse should display a fault. My Fluke gives an "OL" in this case for overload.
There are fuses in these current circuits. Mine have 440mA or 1A fast-blo and I assume yours will have one, too, although I don't know for sure what it's rated. It could be smaller to match the ranges.
So my Klein when I set it to μA range with that 18.37mA (18,370 μA) load on a 4000 μA range goes to OL, too. So yep, handles that correctly. ✔️